TVOC Technical Specifications
|
Sensor (PID) |
10.6 eV lamp with Ion Science patented fence electrode technology |
Intrinsically safe
approvals |
II 2 G Baseefa 03ATEX0742 EEx ia II T4 (-20 ¡ÆC < Ta < 60)
IECEx BAS 04.0033 (world standard) |
Ingress protection rating |
IP65 - dependent on cable glands fitted
Sensor - IP54 |
Power |
5-28 VDC Max 70 mA |
Output |
4 - 20 mA (requires an 8 -35 VDC power supply) |
Range |
User selectable 0.1 - 100 ppm or 1 - 1000 ppm |
Sampling |
Diffusion (can be pumped if required using an adaptor) |
Display |
7 segment, 4 digit LCD, 4 colour LEDs |
Response |
Sensor - T90 < 5 sec
Detector - 1 minute as standard* |
Accuracy |
0 - 100 ppm + 2%
100 - 1000 ppm + 5% |
Linearity |
0 - 100 ppm >98% |
Calibration |
Magnetically accessed
100 ppm Isobutylene via calibration tool kit accessory |
Temperature |
-20 to 50 ¡ÆC (-4 to 122 ¡ÆF) |
Weight |
Instrument 1.3 kg (2.9 lb)
Packed 1.47 kg (3.2 lb) |
Dimensions |
124 x 181 x 73 mm (4.9 x 7.1 x 2.9") |
All specifications are against isobutylene calibration at 20 ¡Æ C, 90% RH and up to 100 ppm in air